Show us why the public should always call a land surveyor

Where are you surveying?


🌐 Show Forums for All Locations
Land Surveyor Community

Replies

  • Land Surveyor

    Over what distance are you measuring?

    Certainly check your prism constant but there are many other possibilities.

    1. What measurement mode are you using? Fine will give you +/-(2mm + 2 ppm) accuracy. The other two modes are +/-(5mm + 2ppm), so if you aren't using Fine, 4mm is within the tolerance of the instrument. Likewise if you are measuring over 1000m, expect no better than +/-4mm on Fine.
    2. What about temperature correction? Over longer distances it can be significant. See page 175 of your manual. For instance at 500m, 25C to 30C is about 6 ppm!
    3. When was the last time the instrument was calibrated?
    4. When was the last time the laser you used to check the distance with was calibrated?
    5. What's the expected PPM error for both the instrument and the laser? Does 4mm fall within the expected error? I'm betting it does based on what I've read in your manual.
    6. I assume you are measuring to a prism mounted on a tribrach; set it up over a fixed spot with just the post showing (no prism) and have someone point out the nail (or whatever) and ensure the top middle of the post is directly over the spot on the ground. Now pick up the tripod, rotate 90 degrees, level it and repeat. Next, remove the post and the instrument from their tribrachs, leaving the tripods untouched. Swap the positions, check, rotate 90 and check again. This is an old school method  but really what you are trying to do is ensure your laser / optical plummets are pointing correctly 

    Hope this helps

    • Land Surveyor

      Hi David,

      I used in the fine mode / put the Temperature/Pressure / the baseline is a short distance in this case is 10mts.

      The prism  is in a tripod with tribrach with adapter and the total station is in a collimator( Photo of the collimator).

      I believe that problem is in the instrument, needs to be calibrated.

      In the country that I work, I think there is no Technical Assistance to Leica.

      I have counseled the client to go to a Leica representative.

      Thank you all!1201387520?profile=original?width=721

    • Not a Surveyor

      I'm going back a while now but when we used to use the main Leica prisms for other instruments we used to set the prism constant at 34.4 for zero error.

      Will.

  • Hi Paulo, You said you have a manual,then look on page 25, But here is what to do press menu  for more than 2 sec. You will see Config, Menu,  on #1 you will see  ppm/mm, there is where you can set your prism

    constant, this is in detail on page 91 ppm / page 92 prism mm., Back to config.Menu # 2 Is CALIB instrument error,this is where to check and correct the Gun  1 is V-index and 2 is H Collim,  In this Config. Menu ,you will find all of the settings1. ppm /mm 2. Calib 3.Contrast 4.rec data 5. units 6.time /date 7.on /off.

     You also have a Main Menu with a tree under Menu and Prog menu with a tree.

    hope this helps.

  • Land Surveyor
    Set up with a different instrument and measure an unknown distance between two random points that you have just set...measure 3 times recording each measurement...set up with your "bad instrument" and measure the same way...if the match or are extremely close then you have no "bad instrument" and there is an issue with your original field procedures or prism constants.
  • One possibility is a prism/instrument constant issue.

    Leica instruments expect a prism with 34mm constant, and many prisms are 30mm; which would result in the 4mm that you refer to.

    Something to look into.

    Also, ensure that whatever value you are treating as the correct value is actually correct. If something was established badly there may be no problem with your measurement at all.

    • Land Surveyor

      Hi Timothy,

      The CP is correct.

      I have the same error (4mm) when use laser.

      For example. I did a baseline with 10mts, with diferente total Station everything is ok with and without prism. With this TCR805 I have 4mm.

      Resume the equipment have 4mm error, I just want know if the operator can correct this error.

      I do not know very well Leica equipments.I already read the operator manual and did not see any function to correct EDM. 

This reply was deleted.

Answer As your Facebook Page

Global Surveyor Forum

Latest Discussions by Category

How to Utilize Surveyor Forums

How to use Forums
Our forums on Land Surveyors United are here to be used as much for finding help with problems in the field as the are for you to express your opinions on anything that has to do with land surveying in general. Feel free to share anything that is on your mind, as long as it isn't meant to damage another member's reputation. Please keep it clean and help insure that everyone has the opportunity to enjoy the benefits of being part of a community that grows together.

We are committed to allowing freedom of expression for all of our members, and that includes maintaining a safe space for people with opposing views to express themselves. We get posts from all over the country and even the globe, so needless to say, people come with different viewpoints on lad surveying practices and processes. We see this diversity and variety as a real strength-- dialogue and debate are an integral part of the educational process, as well as an important tool in exploring different sides of complex issues.

All Community Hubs inside the community have their own forum for asking specific questions to other surveyors, by location, equipment type, etc.

Topics by Tags

Monthly Archives

Global Forums

Add Your Expertise

🌐 Explore Forums Based on Location

Blog Topics by Tags

Monthly Archives